Off Set Temperature of Furnace during analysis
Posted: Fri Nov 17, 2023 2:05 am
by kamran_chemist
How to incorporate deviation -9.2C in furnace, while its measured temperature is 909.2C and Display temperature is 900. if there is no option of offset in controller. What will be the set temperature to achieve 900C.
Re: Off Set Temperature of Furnace during analysis
Posted: Fri Nov 17, 2023 6:43 am
by ChenBeier
890,8°C Display,
measured 900 ° C
But confirm with measurement.
Re: Off Set Temperature of Furnace during analysis
Posted: Fri Nov 17, 2023 10:15 pm
by kamran_chemist
Dear All & ChenBeier
See attachment and describe?
Re: Off Set Temperature of Furnace during analysis
Posted: Sat Nov 18, 2023 3:09 am
by ChenBeier
In this case I would draw a curve A to X. And read at 900° C the X value for A.
Strange is that in beginning A is lower as X and later its the opposit. Its not linear.