Off Set Temperature of Furnace during analysis

Post by kamran_chemist »

How to incorporate deviation -9.2C in furnace, while its measured temperature is 909.2C and Display temperature is 900. if there is no option of offset in controller. What will be the set temperature to achieve 900C.
User avatar
ChenBeier
Distinguished Member
Distinguished Member
Posts: 1565
Joined: Wed Sep 27, 2017 7:25 am
Location: Berlin, Germany

Re: Off Set Temperature of Furnace during analysis

Post by ChenBeier »

890,8°C Display,
measured 900 ° C
But confirm with measurement.
kamran_chemist
Newbie
Newbie
Posts: 4
Joined: Tue Apr 06, 2021 1:11 am

Re: Off Set Temperature of Furnace during analysis

Post by kamran_chemist »

Dear All & ChenBeier

See attachment and describe?
Attachments
Calibration Certificate Data
Calibration Certificate Data
CALIBRATION.jpeg (113.12 KiB) Viewed 2678 times
User avatar
ChenBeier
Distinguished Member
Distinguished Member
Posts: 1565
Joined: Wed Sep 27, 2017 7:25 am
Location: Berlin, Germany

Re: Off Set Temperature of Furnace during analysis

Post by ChenBeier »

In this case I would draw a curve A to X. And read at 900° C the X value for A.
Strange is that in beginning A is lower as X and later its the opposit. Its not linear.