Off Set Temperature of Furnace during analysis
How to incorporate deviation -9.2C in furnace, while its measured temperature is 909.2C and Display temperature is 900. if there is no option of offset in controller. What will be the set temperature to achieve 900C.
- ChenBeier
- Distinguished Member
- Posts: 1425
- Joined: Wed Sep 27, 2017 7:25 am
- Location: Berlin, Germany
Re: Off Set Temperature of Furnace during analysis
890,8°C Display,
measured 900 ° C
But confirm with measurement.
measured 900 ° C
But confirm with measurement.
-
- Newbie
- Posts: 4
- Joined: Tue Apr 06, 2021 1:11 am
Re: Off Set Temperature of Furnace during analysis
Dear All & ChenBeier
See attachment and describe?
See attachment and describe?
- Attachments
-
- Calibration Certificate Data
- CALIBRATION.jpeg (113.12 KiB) Viewed 87 times
- ChenBeier
- Distinguished Member
- Posts: 1425
- Joined: Wed Sep 27, 2017 7:25 am
- Location: Berlin, Germany
Re: Off Set Temperature of Furnace during analysis
In this case I would draw a curve A to X. And read at 900° C the X value for A.
Strange is that in beginning A is lower as X and later its the opposit. Its not linear.